As I have said in reply to Leonid, I don't know why we have such a large 
difference in our experimental setup. It may be that our 0.02 slits are 
misaligned -- I will have to check this.

The question of what is meant by signal to noise ratio in connection with 
powder diffraction is one which I have been trying to find out about. You see 
this term used quite often in papers and in general discussion, but I have 
still to discover how one works it out in a properly defined way with regards 
to powder diffraction. It is well defined in radio frequency signals.

One could for example take, as you have said the peak height minus the 
background and divide by the background as one measure, a value which would 
tend to zero as the peak becomes smaller. This definition would also imply that 
increasing measurement time has no effect on signal to noise ratio, since both 
peak and background would simply scale up with time, in which case we might as 
well measure for negligible amount of time.

Or else one could take the peak intensity divided by the square root of the 
background: this at least would improve with measurement time.

For instance suppose we have a peak above background of 1 counts and a 
background of 1000 counts, this would give a signal to noise ratio of roughly 
322. If we measure ten times longer, the peak intensity becomes 10 and the 
background becomes 1, giving a signal to noise ratio of 1000, an 
improvement!
 So my question remains: what is the definition of signal to noise ratio that 
is accepted for powder diffraction?
 

Mike

**

-Original Message-
From: Van der Lee [mailto:[EMAIL PROTECTED] 
Sent: 19 February 2008 07:39
To: rietveld_l@ill.fr
Subject: Re: advice on new powder diffractometer

After having purchased the set of 0.02 soller slits in Dec. 03 I have done some 
comparison tests on the (111) silicon peak under identical conditions except 
for the change of the primary and secondary sollers:
   background Bpeak height PH
0.0295 8580
0.04   32018900
negliable change on the FWHM

so not an intensity drop with a factor of 25, neither 4, but only slightly 
larger than 2.

How do I define signal/background ratio? There is not much scientific in it, it 
serves only to compare qualitatively these kind of 'optical' 
elements; it should be something like (PH-B)/B or more generally Bragg 
scattering intensity to non-Bragg-scattering intensity.

Arie


 Dear Mike,
 
 Normally, changing sollers must not influence the signal/background 
 ratio. Wider sollers, however, make the primary beam wider and if the 
 sample diameter is small then a parasitic scattering from the sample 
 holder edges may appear.
 
 I am really surprised that moving from 0.04 Soller slits to 0.02 you 
 got 25 times intensity reduction. When I change the primary soller 
 from
 0.04 to 0.02 the intensity drops ~2 times, so if you changed both the 
 primary and the secondary sollers the intensity should decrease ~4 
 times, but not 25 times.
 
 Leonid Solovyov
 
 How do you define signal to noise in powder diffraction? I have seen  
 this term used several times, but I have not found a definition so
 far
  with regard to powder diffraction per se. 
 I have just done two runs on a Panalytical one with 0.04 soller slits  
 and one with 0.02 (both with a CuKa1 premonochromator) both for
 about 9
  hours. The strongest peak for the 0.02 case is 2700 counts, half
 width
  0.08 degrees and a background of 25 counts. The same peak with the
 0.04
  slits  has 7 counts, half width 0.11 degrees and a background of  
 700 counts.


 Mike Glazer
 
 
 
   
 __
 __
 Be a better friend, newshound, and
 know-it-all with Yahoo! Mobile.  Try it now.  
 http://mobile.yahoo.com/;_ylt=Ahu06i62sR8HDtDypao8Wcj9tAcJ
 
 


--
***
A. van der Lee
Institut Européen des Membranes
CNRS - UMR 5635
Université de Montpellier II - Case Courrier 047 Place E. Bataillon
34095 MONTPELLIER Cedex 5 - FRANCE

Tel :  33 (0) 4 67 14 91 35
Fax : 33 (0) 4 67 14 91 19

Website X-ray scattering facility ICG/IEM:
http://www.iemm.univ-montp2.fr/xrayweb/main_uk.html





RE: advice on new powder diffractometer

2008-02-19 Thread Michael Glazer
 
Leonid
I don't know why we have such a large difference, but for some reason we
do. I will send you privately the two diagrams.
Mike

-Original Message-
From: Leonid Solovyov [mailto:[EMAIL PROTECTED] 
Sent: 19 February 2008 03:56
To: rietveld_l@ill.fr
Subject: RE: advice on new powder diffractometer

Dear Mike,

Normally, changing sollers must not influence the signal/background
ratio. Wider sollers, however, make the primary beam wider and if the
sample diameter is small then a parasitic scattering from the sample
holder edges may appear.

I am really surprised that moving from 0.04 Soller slits to 0.02 you got
25 times intensity reduction. When I change the primary soller from
0.04 to 0.02 the intensity drops ~2 times, so if you changed both the
primary and the secondary sollers the intensity should decrease ~4
times, but not 25 times.

Leonid Solovyov



Re: advice on new powder diffractometer

2008-02-19 Thread Van der Lee

Michael Glazer wrote the following on 19/02/2008 09:42:


Or else one could take the peak intensity divided by the square root of the 
background: this at least would improve with measurement time.

For instance suppose we have a peak above background of 1 counts and a 
background of 1000 counts, this would give a signal to noise ratio of roughly 
322. If we measure ten times longer, the peak intensity becomes 10 and the 
background becomes 1, giving a signal to noise ratio of 1000, an 
improvement!
 So my question remains: what is the definition of signal to noise ratio that 
is accepted for powder diffraction?
 

This at least coincides with data collection practice in single crystal 
diffraction: there it is said that in order to improve the 'signal to 
noise' ratio with a factor of two you have to count 4 times longer, or, 
alternatively, double the generator power.


Arie




RE: advice on new powder diffractometer

2008-02-19 Thread Alan Hewat
 So my question remains: what is the definition of signal to noise ratio
 that is accepted for powder diffraction?

Why does it matter? A higher Bragg/Background ratio does not necessarily
mean better data if counting statistics are poor. Exaggerating slightly
:-) consider a single peak with a ratio of 100/1 compared to a peak with a
ratio 1/1000. The second measurement will give the lowest error, not
the first which has a much higher signal/noise. And you measure lots of
points on a slowly varying background, so you have a much better estimate
of background than the normal error of a single point. Please don't
encourage people to simply maximise signal/noise.

Similarly, low profile R-factor's can be obtained with low resolution data
and high background. That does not mean that low resolution data produces
smaller errors in structural parameters.

I worry about people treating measurement and refinement as black boxes
with simplified measures of quality such as R-factors, signal-to-noise
etc. You have to look at the physical reality of the model and the
estimated errors in its parameters, while not cheating by removing data
that doesn't fit for unknown reasons, adding too much a priori
information such as constraints, or throwing in extra garbage parameters
to improve the R-factors.
__
Dr Alan Hewat, NeutronOptics, Grenoble, FRANCE
[EMAIL PROTECTED] +33.476.98.41.68
http://www.NeutronOptics.com/
__



RE: advice on new powder diffractometer

2008-02-19 Thread Leonid Solovyov
 It seems to me that if there is no internationally accepted
definition
 of signal to noise ratio in powder diffraction, then let me suggest
 that
 this forum might be a way to define it once and for all. How about
 this?
   SNR = (P-B)/SQRT(B). 

 Of course this would only give a  value for a single peak. I suppose
 one could also define a global SNR as a sum over all 
 Global SNR = sum over peaks (P-B)/SQRT(B).

 Does this make sense?
 Mike

A more overall value would be a normalized sum of the ratios
(Ip-Ib)/sqrt(Ib), where Ip and Ib are the profile intensities of the
pattern and the background for each datapoint.

Leonid



  

Looking for last minute shopping deals?  
Find them fast with Yahoo! Search.  
http://tools.search.yahoo.com/newsearch/category.php?category=shopping


RE: advice on new powder diffractometer

2008-02-19 Thread gregor
I think that care has to be taken when defining B=noise. In many cases, B 
contains some amorphous material scattering, air scattering, TDS, etc which 
one might account for using a proper model. So they will scale with time 
while the error part of noise will not.

miguel


On 19 Feb 2008 at 9:21, Michael Glazer wrote:

 It seems to me that if there is no internationally accepted definition
 of signal to noise ratio in powder diffraction, then let me suggest that
 this forum might be a way to define it once and for all. How about this?
   SNR = (P-B)/SQRT(B). 
 
 Of course this would only give a  value for a single peak. I suppose one
 could also define a global SNR as a sum over all 
 Global SNR = sum over peaks (P-B)/SQRT(B).
 
 Does this make sense?
 Mike
 
 -Original Message-
 From: Van der Lee [mailto:[EMAIL PROTECTED] 
 Sent: 19 February 2008 09:09
 To: rietveld_l@ill.fr
 Subject: Re: advice on new powder diffractometer
 
 Michael Glazer wrote the following on 19/02/2008 09:42:
  
  Or else one could take the peak intensity divided by the square root
 of the background: this at least would improve with measurement time.
  
  For instance suppose we have a peak above background of 1 counts
 and a background of 1000 counts, this would give a signal to noise ratio
 of roughly 322. If we measure ten times longer, the peak intensity
 becomes 10 and the background becomes 1, giving a signal to
 noise ratio of 1000, an improvement!
   So my question remains: what is the definition of signal to noise
 ratio that is accepted for powder diffraction?
   
 
 This at least coincides with data collection practice in single crystal
 diffraction: there it is said that in order to improve the 'signal to
 noise' ratio with a factor of two you have to count 4 times longer, or,
 alternatively, double the generator power.
 
 Arie
 
 
 

--
Miguel Gregorkiewitz
Dip Scienze della Terra, Università
via Laterina 8, I-53100 Siena, Europe
fon +39'0577'233810 fax 233938
email [EMAIL PROTECTED]





RE: advice on new powder diffractometer

2008-02-19 Thread gregor
in general, I made the observation that intensity may drop more than 
expected from theory due to alignment problems. Alignment becomes 
increasingly critical for smaller divergences as we normally have to sintonize 
2 slits, one in the incident and one in the diffracted beam. With low 
divergences, a small twist of the soller axis around theta or the X-ray beam 
may cut an important portion of intensity.

BTW, in our old Philips with 2 Soller 0.04 rad slits and sec beam graphite 
monochromator, using 1º div slits and a 0.2 mm receiving slit gives
12550 cps and FWHM=0.119º (including alpha2 which is visible as a 
shoulder) for the 2th=26.6º quartz peak and ~15 cps for the background 
(zero bk Si sample holder) below this peak.

miguel


On 19 Feb 2008 at 8:29, Michael Glazer wrote:

  
 Leonid
 I don't know why we have such a large difference, but for some reason we
 do. I will send you privately the two diagrams.
 Mike
 
 -Original Message-
 From: Leonid Solovyov [mailto:[EMAIL PROTECTED] 
 Sent: 19 February 2008 03:56
 To: rietveld_l@ill.fr
 Subject: RE: advice on new powder diffractometer
 
 Dear Mike,
 
 Normally, changing sollers must not influence the signal/background
 ratio. Wider sollers, however, make the primary beam wider and if the
 sample diameter is small then a parasitic scattering from the sample
 holder edges may appear.
 
 I am really surprised that moving from 0.04 Soller slits to 0.02 you got
 25 times intensity reduction. When I change the primary soller from
 0.04 to 0.02 the intensity drops ~2 times, so if you changed both the
 primary and the secondary sollers the intensity should decrease ~4
 times, but not 25 times.
 
 Leonid Solovyov
 

--
Miguel Gregorkiewitz
Dip Scienze della Terra, Università
via Laterina 8, I-53100 Siena, Europe
fon +39'0577'233810 fax 233938
email [EMAIL PROTECTED]





Antw: RE: advice on new powder diffractometer

2008-02-19 Thread Nicola Doebelin
From my understanding 'noise' is the random signal superimposed to the
background and the peaks. The signal to noise ratio would be peak above
background / random signal, and this should improve with longer
measurement time.

Nic

 Michael Glazer [EMAIL PROTECTED] 02/19/08 9:42 am 
As I have said in reply to Leonid, I don't know why we have such a
large difference in our experimental setup. It may be that our 0.02
slits are misaligned -- I will have to check this.

The question of what is meant by signal to noise ratio in connection
with powder diffraction is one which I have been trying to find out
about. You see this term used quite often in papers and in general
discussion, but I have still to discover how one works it out in a
properly defined way with regards to powder diffraction. It is well
defined in radio frequency signals.

One could for example take, as you have said the peak height minus the
background and divide by the background as one measure, a value which
would tend to zero as the peak becomes smaller. This definition would
also imply that increasing measurement time has no effect on signal to
noise ratio, since both peak and background would simply scale up with
time, in which case we might as well measure for negligible amount of
time.

Or else one could take the peak intensity divided by the square root of
the background: this at least would improve with measurement time.

For instance suppose we have a peak above background of 1 counts
and a background of 1000 counts, this would give a signal to noise ratio
of roughly 322. If we measure ten times longer, the peak intensity
becomes 10 and the background becomes 1, giving a signal to
noise ratio of 1000, an improvement!
 So my question remains: what is the definition of signal to noise
ratio that is accepted for powder diffraction?
 

Mike

**

-Original Message-
From: Van der Lee [mailto:[EMAIL PROTECTED] 
Sent: 19 February 2008 07:39
To: rietveld_l@ill.fr 
Subject: Re: advice on new powder diffractometer

After having purchased the set of 0.02 soller slits in Dec. 03 I have
done some comparison tests on the (111) silicon peak under identical
conditions except for the change of the primary and secondary sollers:
   background Bpeak height PH
0.0295 8580
0.04   32018900
negliable change on the FWHM

so not an intensity drop with a factor of 25, neither 4, but only
slightly larger than 2.

How do I define signal/background ratio? There is not much scientific
in it, it serves only to compare qualitatively these kind of 'optical' 
elements; it should be something like (PH-B)/B or more generally Bragg
scattering intensity to non-Bragg-scattering intensity.

Arie


 Dear Mike,
 
 Normally, changing sollers must not influence the signal/background 
 ratio. Wider sollers, however, make the primary beam wider and if the

 sample diameter is small then a parasitic scattering from the sample

 holder edges may appear.
 
 I am really surprised that moving from 0.04 Soller slits to 0.02 you

 got 25 times intensity reduction. When I change the primary soller 
 from
 0.04 to 0.02 the intensity drops ~2 times, so if you changed both the

 primary and the secondary sollers the intensity should decrease ~4 
 times, but not 25 times.
 
 Leonid Solovyov
 
 How do you define signal to noise in powder diffraction? I have seen
 
 this term used several times, but I have not found a definition so
 far
  with regard to powder diffraction per se. 
 I have just done two runs on a Panalytical one with 0.04 soller
slits  
 and one with 0.02 (both with a CuKa1 premonochromator) both for
 about 9
  hours. The strongest peak for the 0.02 case is 2700 counts, half
 width
  0.08 degrees and a background of 25 counts. The same peak with the
 0.04
  slits  has 7 counts, half width 0.11 degrees and a background
of  
 700 counts.


 Mike Glazer
 
 
 
   

__
 __
 Be a better friend, newshound, and
 know-it-all with Yahoo! Mobile.  Try it now.  
 http://mobile.yahoo.com/;_ylt=Ahu06i62sR8HDtDypao8Wcj9tAcJ 
 
 


--
***
A. van der Lee
Institut Européen des Membranes
CNRS - UMR 5635
Université de Montpellier II - Case Courrier 047 Place E. Bataillon
34095 MONTPELLIER Cedex 5 - FRANCE

Tel :  33 (0) 4 67 14 91 35
Fax : 33 (0) 4 67 14 91 19

Website X-ray scattering facility ICG/IEM:
http://www.iemm.univ-montp2.fr/xrayweb/main_uk.html 





RE: advice on new powder diffractometer

2008-02-19 Thread Michael Glazer
Alan
I absolutely agree with you regarding Rietveld refinement. The same
argument is true for adjusting peak shapes to get the best fit, playing
around with different peak formulae. Often such tweaking, while it makes
the observed and calculated fit look nicer, has little effect on the
atomic positions, which in the end is what one is trying to derive. One
should not lose sight of the model.

But my question about defining SNR is more to do with the use of powder
data in general, and especially as used in industry or forensics. For
example distinguishing in a legal sense between two materials may depend
on which of two powder patterns has the higher SNR. I keep seeing this
term used in papers and books but never defined.
Mike

-Original Message-
From: Alan Hewat [mailto:[EMAIL PROTECTED] 
Sent: 19 February 2008 10:10
To: rietveld_l@ill.fr
Subject: RE: advice on new powder diffractometer

 So my question remains: what is the definition of signal to noise 
 ratio that is accepted for powder diffraction?

Why does it matter? A higher Bragg/Background ratio does not necessarily
mean better data if counting statistics are poor. Exaggerating slightly
:-) consider a single peak with a ratio of 100/1 compared to a peak with
a ratio 1/1000. The second measurement will give the lowest error,
not the first which has a much higher signal/noise. And you measure lots
of points on a slowly varying background, so you have a much better
estimate of background than the normal error of a single point. Please
don't encourage people to simply maximise signal/noise.

Similarly, low profile R-factor's can be obtained with low resolution
data and high background. That does not mean that low resolution data
produces smaller errors in structural parameters.

I worry about people treating measurement and refinement as black boxes
with simplified measures of quality such as R-factors, signal-to-noise
etc. You have to look at the physical reality of the model and the
estimated errors in its parameters, while not cheating by removing data
that doesn't fit for unknown reasons, adding too much a priori
information such as constraints, or throwing in extra garbage parameters
to improve the R-factors.
__
Dr Alan Hewat, NeutronOptics, Grenoble, FRANCE
[EMAIL PROTECTED] +33.476.98.41.68
http://www.NeutronOptics.com/
__




RE: advice on new powder diffractometer

2008-02-19 Thread Lubomir Smrcok

Gentlemen,
I really do not want to interfere too much, but you should (probably) 
start with the definition of noise and signal. In the theory of 
signals both terms have their well-defined meaning. So you can either 
accept the definition used by the signal community and identify what is 
what in your powder patterns or develop your own definition and put it on 
a good statistical base.

My apologies once more,
Lubo


On Tue, 19 Feb 2008, Michael Glazer wrote:


Alan
I absolutely agree with you regarding Rietveld refinement. The same
argument is true for adjusting peak shapes to get the best fit, playing
around with different peak formulae. Often such tweaking, while it makes
the observed and calculated fit look nicer, has little effect on the
atomic positions, which in the end is what one is trying to derive. One
should not lose sight of the model.

But my question about defining SNR is more to do with the use of powder
data in general, and especially as used in industry or forensics. For
example distinguishing in a legal sense between two materials may depend
on which of two powder patterns has the higher SNR. I keep seeing this
term used in papers and books but never defined.
Mike

-Original Message-
From: Alan Hewat [mailto:[EMAIL PROTECTED]
Sent: 19 February 2008 10:10
To: rietveld_l@ill.fr
Subject: RE: advice on new powder diffractometer


So my question remains: what is the definition of signal to noise
ratio that is accepted for powder diffraction?


Why does it matter? A higher Bragg/Background ratio does not necessarily
mean better data if counting statistics are poor. Exaggerating slightly
:-) consider a single peak with a ratio of 100/1 compared to a peak with
a ratio 1/1000. The second measurement will give the lowest error,
not the first which has a much higher signal/noise. And you measure lots
of points on a slowly varying background, so you have a much better
estimate of background than the normal error of a single point. Please
don't encourage people to simply maximise signal/noise.

Similarly, low profile R-factor's can be obtained with low resolution
data and high background. That does not mean that low resolution data
produces smaller errors in structural parameters.

I worry about people treating measurement and refinement as black boxes
with simplified measures of quality such as R-factors, signal-to-noise
etc. You have to look at the physical reality of the model and the
estimated errors in its parameters, while not cheating by removing data
that doesn't fit for unknown reasons, adding too much a priori
information such as constraints, or throwing in extra garbage parameters
to improve the R-factors.
__
Dr Alan Hewat, NeutronOptics, Grenoble, FRANCE
[EMAIL PROTECTED] +33.476.98.41.68
   http://www.NeutronOptics.com/
__





RE: advice on new powder diffractometer

2008-02-19 Thread Kurt Leinenweber


Hi Bob,

The band of delt/sig is a little wider than that, but I think the real
problem is that they sent me the data in cps.  I am awaiting the data in
counts now.  Thanks to you and others for pointing this out.

However, in my view, having the data in cps should simply go into the
histogram scale factor and should not affect the fit.  What is wrong
with my logic?

Thanks,

- Kurt



-Original Message-
From: Von Dreele, Robert B. [mailto:[EMAIL PROTECTED] 
Sent: Friday, February 15, 2008 6:37 AM
To: rietveld_l@ill.fr
Subject: RE: advice on new powder diffractometer

Hi Kurt,
If you plot the results in POWPLOT  do the statistical analysis what
does the delt/sig vs 2-theta plot look like? If weights are right then
the band of delt/sig ought to be roughly 2-3 sig wide. Is your data in
cpm? Some instruments make that by default. That will mess up the
R-factors.
Bob






RE: advice on new powder diffractometer

2008-02-19 Thread Von Dreele, Robert B.
Kurt (and others),
Intensity in cps isn't a problem if the conversion is also applied to the esd 
and both intensity and esd is output to be read by GSAS. The problem is that 
the instrument software doesn't give you the esds at all but just a table (in 
the old 10 column format) of the intensities as cps. GSAS then assumes 
these are counts and that esd = sqrt(counts) according to the usual Gaussian 
statistics. So chi**2 will be off by the reciprocal of the count time per step, 
i.e. too small. BTW this does not affect the resulting parameters (or esds) one 
bit - it only makes the chi**2 wrong. BTW**2: Poisson statistics only comes 
into play when the count is less than ~20 so that zero is within (say) 4-5 
sigma of the count. Above that Gaussian statistics ~ Poisson statistics. So if 
all points in the scan have 20 counts (including background) then the 
assumptions about Gaussian statistics for the counts are reasonable. BTW**3: If 
the background was say 20 counts with a sig ~4-5 then the R-!
 value for that background will be ~20%. So if your pattern has very few peaks 
(say only 10% of the points) then R-values can easily be much larger than 10%. 
Following this logic a bit further, a pattern with a higher background will 
give lower R-values so IMHO comparing R-values from one experiment to another 
is a particularly pointless exercise. Ditto for asking how good does my 
R-factor have to be? because it just depends.
Bob


-Original Message-
From: Kurt Leinenweber [mailto:[EMAIL PROTECTED]
Sent: Tue 2/19/2008 10:37 AM
To: rietveld_l@ill.fr
Subject: RE: advice on new powder diffractometer
 


Hi Bob,

The band of delt/sig is a little wider than that, but I think the real
problem is that they sent me the data in cps.  I am awaiting the data in
counts now.  Thanks to you and others for pointing this out.

However, in my view, having the data in cps should simply go into the
histogram scale factor and should not affect the fit.  What is wrong
with my logic?

Thanks,

- Kurt



-Original Message-
From: Von Dreele, Robert B. [mailto:[EMAIL PROTECTED] 
Sent: Friday, February 15, 2008 6:37 AM
To: rietveld_l@ill.fr
Subject: RE: advice on new powder diffractometer

Hi Kurt,
If you plot the results in POWPLOT  do the statistical analysis what
does the delt/sig vs 2-theta plot look like? If weights are right then
the band of delt/sig ought to be roughly 2-3 sig wide. Is your data in
cpm? Some instruments make that by default. That will mess up the
R-factors.
Bob








RE: advice on new powder diffractometer

2008-02-19 Thread Kurt Leinenweber
Hi all,

I never thought of background as noise - to me the noise should be
defined as the esd of the background points.  You could get that by
averaging 10 or 20 background points and getting the standard deviation.
I'm told that for typical counting statistics, that should be close to
the square root of the background, so Mike's idea of using that square
root seems reasonable.

- Kurt

-Original Message-
From: Alan Hewat [mailto:[EMAIL PROTECTED] 
Sent: Tuesday, February 19, 2008 3:10 AM
To: rietveld_l@ill.fr
Subject: RE: advice on new powder diffractometer

 So my question remains: what is the definition of signal to noise
ratio
 that is accepted for powder diffraction?

Why does it matter? A higher Bragg/Background ratio does not necessarily
mean better data if counting statistics are poor. Exaggerating slightly
:-) consider a single peak with a ratio of 100/1 compared to a peak with
a
ratio 1/1000. The second measurement will give the lowest error, not
the first which has a much higher signal/noise. And you measure lots of
points on a slowly varying background, so you have a much better
estimate
of background than the normal error of a single point. Please don't
encourage people to simply maximise signal/noise.

Similarly, low profile R-factor's can be obtained with low resolution
data
and high background. That does not mean that low resolution data
produces
smaller errors in structural parameters.

I worry about people treating measurement and refinement as black boxes
with simplified measures of quality such as R-factors, signal-to-noise
etc. You have to look at the physical reality of the model and the
estimated errors in its parameters, while not cheating by removing data
that doesn't fit for unknown reasons, adding too much a priori
information such as constraints, or throwing in extra garbage parameters
to improve the R-factors.
__
Dr Alan Hewat, NeutronOptics, Grenoble, FRANCE
[EMAIL PROTECTED] +33.476.98.41.68
http://www.NeutronOptics.com/
__




RE: advice on new powder diffractometer

2008-02-19 Thread Worlton, Thomas G.
I think the problem with creating a new definition of signal to noise is
that this is an attempt to make something that corresponds to the
quality of the measurement, while the true signal to noise is only a
measure of the quality of the signal.  This can be used for estimating
the time needed for a measurement to obtain the desired statistics, but
should not be used as a way to judge the quality of a completed
measurement. 

As far as taking the sum of the squares, it is my understanding that the
signal to noise is intended to be applied to a single mode, so the true
signal-to-noise could vary widely for different modes.  As I mentioned,
it can be a very useful predictor of the time needed for the
measurement, but probably shouldn't be used in discussing the quality of
the measurement.  If you are interested in a peak other than the one for
the highest mode, you should use that peak when estimating the time
needed for a measurement.

Another link is http://en.wikipedia.org/wiki/Signal-to-noise_ratio

Tom Worlton

-Original Message-
From: Leonid Solovyov [mailto:[EMAIL PROTECTED] 
Sent: Tuesday, February 19, 2008 3:51 AM
To: rietveld_l@ill.fr
Subject: RE: advice on new powder diffractometer

 It seems to me that if there is no internationally accepted
definition
 of signal to noise ratio in powder diffraction, then let me suggest  
that  this forum might be a way to define it once and for all. How 
about  this?
   SNR = (P-B)/SQRT(B). 

 Of course this would only give a  value for a single peak. I suppose 
 one could also define a global SNR as a sum over all Global SNR = sum 
 over peaks (P-B)/SQRT(B).

 Does this make sense?
 Mike

A more overall value would be a normalized sum of the ratios
(Ip-Ib)/sqrt(Ib), where Ip and Ib are the profile intensities of the
pattern and the background for each datapoint.

Leonid



 


Looking for last minute shopping deals?  
Find them fast with Yahoo! Search.
http://tools.search.yahoo.com/newsearch/category.php?category=shopping



RE: advice on new powder diffractometer

2008-02-19 Thread May, Frank
I've watched this thread develop over the past few days.  Apparently there is 
little advice forthcoming about the original inquiry.

Just one person's observation.

Frank May - University of Missouri-St. Louis

winmail.dat

Re: advice on new powder diffractometer

2008-02-18 Thread William Bisson

Yes it is mainly down to the Soller slits, there was a very large thread on
soller slits somewhere in the Rietveld archives about this discussion. I think
the down side of changing the soller slits is a move away from the 
optimum FWHM

that can be obtained?

Quoting Leonid Solovyov [EMAIL PROTECTED]:


Xpert Pros are fantastic machines. Collections times are

significantly

reduced
with their new semi-conductor PSDs. The only snag with the machines

is=20

that the
asymmetry at low angle is very stark - so much so that in effect you=20
are having
to model a multiple diffractometer situation...


In fact, the problem of asymmetry is related to the Soller slit choice
(and to the diffractometer radius), but not to the detector. By default
PANal supplies their machines with Soller slits of 0.04 rad divergence
that give highly asymmetric peaks, but one may choose Sollers of 0.02
or 0.01 rad to reduce the asymmetry. Additionally, the secondary
monochromator for X=92Celerator or PIXcel detectors makes the peaks even
more symmetric as it suppresses the axial divergence.

Leonid Solovyov



 ___=
_
Be a better friend, newshound, and=20
know-it-all with Yahoo! Mobile.  Try it now.  http://mobile.yahoo.com/;_y=
lt=3DAhu06i62sR8HDtDypao8Wcj9tAcJ=20






William Bisson

--
Webmaster and Administrator - CCP14 and BCA
http://www.ccp14.ac.uk
http://bca.chem.ucl.ac.uk/


Re: advice on new powder diffractometer

2008-02-18 Thread Leonid Solovyov
 Yes it is mainly down to the Soller slits, 
 there was a very large thread on
 soller slits somewhere in the Rietveld archives
 about this discussion. I think the down side of
 changing the soller slits is a move away from the
 optimum FWHM that can be obtained?

Changing sollers from 0.04 rad to 0.02 or 0.01 reduces the asymmetry,
the FWHM and the intensity, so the peak shape and the resolution become
more optimal but the intensity is sacrificed.



  

Never miss a thing.  Make Yahoo your home page. 
http://www.yahoo.com/r/hs


Re: advice on new powder diffractometer

2008-02-18 Thread Van der Lee

Leonid Solovyov wrote the following on 18/02/2008 16:27:
Yes it is mainly down to the Soller slits, 
there was a very large thread on

soller slits somewhere in the Rietveld archives
about this discussion. I think the down side of
changing the soller slits is a move away from the
optimum FWHM that can be obtained?


Changing sollers from 0.04 rad to 0.02 or 0.01 reduces the asymmetry,
the FWHM and the intensity, so the peak shape and the resolution become
more optimal but the intensity is sacrificed.




However, the signal to noise ratio becomes better with 0.02/0.01 rad 
sollers compared to 0.04 rad


Arie


--
***
A. van der Lee
Institut Européen des Membranes
CNRS - UMR 5635
Université de Montpellier II - Case Courrier 047
Place E. Bataillon
34095 MONTPELLIER Cedex 5 - FRANCE

Tel :  33 (0) 4 67 14 91 35
Fax : 33 (0) 4 67 14 91 19

Website X-ray scattering facility ICG/IEM:
http://www.iemm.univ-montp2.fr/xrayweb/main_uk.html




RE: advice on new powder diffractometer

2008-02-18 Thread Michael Glazer
How do you define signal to noise in powder diffraction? I have seen this term 
used several times, but I have not found a definition so far with regard to 
powder diffraction per se. 
I have just done two runs on a Panalytical one with 0.04 soller slits and one 
with 0.02 (both with a CuKa1 premonochromator) both for about 9 hours. The 
strongest peak for the 0.02 case is 2700 counts, half width 0.08 degrees and a 
background of 25 counts. The same peak with the 0.04 slits  has 7 counts, 
half width 0.11 degrees and a background of 700 counts.


Mike Glazer

-Original Message-
From: Van der Lee [mailto:[EMAIL PROTECTED] 
Sent: 18 February 2008 16:36
To: rietveld_l@ill.fr
Subject: Re: advice on new powder diffractometer

Leonid Solovyov wrote the following on 18/02/2008 16:27:
 Yes it is mainly down to the Soller slits, there was a very large 
 thread on soller slits somewhere in the Rietveld archives about this 
 discussion. I think the down side of changing the soller slits is a 
 move away from the optimum FWHM that can be obtained?
 
 Changing sollers from 0.04 rad to 0.02 or 0.01 reduces the asymmetry, 
 the FWHM and the intensity, so the peak shape and the resolution 
 become more optimal but the intensity is sacrificed.
 


However, the signal to noise ratio becomes better with 0.02/0.01 rad sollers 
compared to 0.04 rad

Arie


--
***
A. van der Lee
Institut Européen des Membranes
CNRS - UMR 5635
Université de Montpellier II - Case Courrier 047 Place E. Bataillon
34095 MONTPELLIER Cedex 5 - FRANCE

Tel :  33 (0) 4 67 14 91 35
Fax : 33 (0) 4 67 14 91 19

Website X-ray scattering facility ICG/IEM:
http://www.iemm.univ-montp2.fr/xrayweb/main_uk.html





RE: advice on new powder diffractometer

2008-02-18 Thread Lubomir Smrcok

IMHO,
plenty of counter problems have their roots in the past.  When 
the first GM counters were used in powder diffractometry the apparatus 
(both 
electronics  math) was taken directly from the field of \alpha-particles 
physics. 
Because \alpha-particles are rather rare (events), people used binomial or 
POissonian distributions with mean value=variance. This is hardly true for 
normal (Gaussian) distribution  diffraction. Later when the LS moved to 
the field of diffraction, 
weights were automatically set to w=1/var as if the random errors (=noise, 
or better, white noise) had normal distribution. So far so good, 
but  variance was set to 
the mean value as if it were Poissonian distribution. Because the LS are 
derived under the 
assumption of normal distribution of random errors, the procedure must 
give biased estimates of the parameters. Now when the solid state 
detectors are rather fast and efficient it is probably the high time for 
new data collection strategies, e.g. repeated measurements giving better 
estimates of the variances. In this way it could be possible to estimate 
also the noise level. Who knows, BTW :-0

Lubo


On Mon, 18 Feb 2008, Michael Glazer wrote:


How do you define signal to noise in powder diffraction? I have seen this term 
used several times, but I have not found a definition so far with regard to 
powder diffraction per se.
I have just done two runs on a Panalytical one with 0.04 soller slits and one 
with 0.02 (both with a CuKa1 premonochromator) both for about 9 hours. The 
strongest peak for the 0.02 case is 2700 counts, half width 0.08 degrees and a 
background of 25 counts. The same peak with the 0.04 slits  has 7 counts, 
half width 0.11 degrees and a background of 700 counts.


Mike Glazer

-Original Message-
From: Van der Lee [mailto:[EMAIL PROTECTED]
Sent: 18 February 2008 16:36
To: rietveld_l@ill.fr
Subject: Re: advice on new powder diffractometer

Leonid Solovyov wrote the following on 18/02/2008 16:27:

Yes it is mainly down to the Soller slits, there was a very large
thread on soller slits somewhere in the Rietveld archives about this
discussion. I think the down side of changing the soller slits is a
move away from the optimum FWHM that can be obtained?


Changing sollers from 0.04 rad to 0.02 or 0.01 reduces the asymmetry,
the FWHM and the intensity, so the peak shape and the resolution
become more optimal but the intensity is sacrificed.




However, the signal to noise ratio becomes better with 0.02/0.01 rad sollers 
compared to 0.04 rad

Arie


--
***
A. van der Lee
Institut Européen des Membranes
CNRS - UMR 5635
Université de Montpellier II - Case Courrier 047 Place E. Bataillon
34095 MONTPELLIER Cedex 5 - FRANCE

Tel :  33 (0) 4 67 14 91 35
Fax : 33 (0) 4 67 14 91 19

Website X-ray scattering facility ICG/IEM:
http://www.iemm.univ-montp2.fr/xrayweb/main_uk.html





RE: advice on new powder diffractometer

2008-02-18 Thread Alan Hewat
As Mike implies, for a given counting time, you have much better
statistics using the high intensity configuration. This can be true even
when the signal to noise ratio is actually lower with the high intensity
configuration. So signal/noise is not everything.

That said, there is no point in reducing statistical error much below
systematic error. I would try hard to physically model the peak asymmetry
using a higher intensity configuration, and for a given counting time try
to balance statistical error with systematic error, which occurs when the
model cannot completely describe the data however you choose the model
parameters.

Many theoretical arguments about signal/noise forget that the
experimentalist has a finite time to collect his data :-)

Alan.

Michael Glazer said:
 I have just done two runs on a Panalytical one with 0.04 soller slits and
 one with 0.02 (both with a CuKa1 premonochromator) both for about 9 hours.
 The strongest peak for the 0.02 case is 2700 counts, half width 0.08
 degrees and a background of 25 counts. The same peak with the 0.04 slits
 has 7 counts, half width 0.11 degrees and a background of 700 counts.
__
Dr Alan Hewat, NeutronOptics, Grenoble, FRANCE
[EMAIL PROTECTED] +33.476.98.41.68
http://www.NeutronOptics.com/
__



Re: advice on new powder diffractometer

2008-02-18 Thread Jon Wright

Alan Hewat wrote:

Many theoretical arguments about signal/noise forget that the
experimentalist has a finite time to collect his data :-)


The finite time to analyse and publish can also be rate limiting! In 
this respect a high background is a fine addition to any instrument; it 
gives a much lower profile R-factor. This tells us something about the 
value of the profile R factor, I think ;-)


Best,

Jon


Re: advice on new powder diffractometer

2008-02-17 Thread Leonid Solovyov
 Xpert Pros are fantastic machines. Collections times are
significantly
 reduced
 with their new semi-conductor PSDs. The only snag with the machines
is 
 that the
 asymmetry at low angle is very stark - so much so that in effect you 
 are having
 to model a multiple diffractometer situation...

In fact, the problem of asymmetry is related to the Soller slit choice
(and to the diffractometer radius), but not to the detector. By default
PANal supplies their machines with Soller slits of 0.04 rad divergence
that give highly asymmetric peaks, but one may choose Sollers of 0.02
or 0.01 rad to reduce the asymmetry. Additionally, the secondary
monochromator for X’Celerator or PIXcel detectors makes the peaks even
more symmetric as it suppresses the axial divergence.

Leonid Solovyov



  

Be a better friend, newshound, and 
know-it-all with Yahoo! Mobile.  Try it now.  
http://mobile.yahoo.com/;_ylt=Ahu06i62sR8HDtDypao8Wcj9tAcJ 



Re: advice on new powder diffractometer

2008-02-16 Thread William Bisson

Dear Kurt,

Xpert Pros are fantastic machines. Collections times are significantly reduced
with their new semi-conductor PSDs. The only snag with the machines is 
that the
asymmetry at low angle is very stark - so much so that in effect you 
are having

to model a multiple diffractometer situation. If you try and us fundamental
parameters to fit the date you run up against a commercial secret as to the
separation between each of the semi-conducting strips in the detector.

I don't know as yet, any Rietveld program that have profile functions that can
adequately model the low angle data from the new X'pert pro machines. 
If anyone

does I'll be glad to know, and from a fundamental parameters approach what the
separation and/or width of the semi-conductor strips are?

Regards
William Bisson



Quoting Kurt Leinenweber [EMAIL PROTECTED]:


Hi All,

I am attempting to assist others to choose a new powder x-ray
diffractometer.  We are currently looking at a Panalytical X'Pert Pro in
alpha-1 configuration, and I would like to know if anyone has any
experience with this for Rietveld refinements?  For other reasons we
would like to keep the sample flat, but then we could not use the
alpha-1 configuration, so instead we would consider a hybrid
monochromator - I would like to ask if this monochromator makes it
impossible to do Rietveld?  Also, any other insights and suggestions for
systems that are competitive with these would be welcome.  We are at an
early stage and have time to think about this, will be looking at other
possibilities, and I would appreciate any suggestions.

Incidentally, I received some data on quartz (supplied by me) from the
alpha-1, and refined it using GSAS.  The refined structural and thermal
parameters are fine, but I get strange r-factors, with wRp of 25% and
crystallographic R also near 25%, chi**2 of 0.5.  Does anyone know why
this might happen?  With my old Siemens D5000 I easily get down to 10%
or better in the r-factors right away, so something obvious must be
wrong.

Thank you,

- Kurt Leinenweber










William Bisson

--
Webmaster and Administrator - CCP14 and BCA
http://www.ccp14.ac.uk
http://bca.chem.ucl.ac.uk/


Re: advice on new powder diffractometer

2008-02-15 Thread Jose marcos Sasaki

Dear Kurt,
The only thing that I remember is to change cos(2theta) of monochromator 
in polarization factor or precisely calculate the polarization factor 
for n-reflection in the monochromator.

Regards
Sasaki


Kurt Leinenweber escreveu:

Hi All,

I am attempting to assist others to choose a new powder x-ray
diffractometer.  We are currently looking at a Panalytical X'Pert Pro in
alpha-1 configuration, and I would like to know if anyone has any
experience with this for Rietveld refinements?  For other reasons we
would like to keep the sample flat, but then we could not use the
alpha-1 configuration, so instead we would consider a hybrid
monochromator - I would like to ask if this monochromator makes it
impossible to do Rietveld?  Also, any other insights and suggestions for
systems that are competitive with these would be welcome.  We are at an
early stage and have time to think about this, will be looking at other
possibilities, and I would appreciate any suggestions.

Incidentally, I received some data on quartz (supplied by me) from the
alpha-1, and refined it using GSAS.  The refined structural and thermal
parameters are fine, but I get strange r-factors, with wRp of 25% and
crystallographic R also near 25%, chi**2 of 0.5.  Does anyone know why
this might happen?  With my old Siemens D5000 I easily get down to 10%
or better in the r-factors right away, so something obvious must be
wrong.

Thank you,

- Kurt Leinenweber





  



--
___
Dr. José Marcos Sasaki (professor associado)
Departamento de Física
Universidade Federal do Ceará - UFC
Campus do Pici CEP 60455-760
Caixa Postal 6030
Fortaleza - Ceará - Brazil
Telefone: 55 (0xx85) 3366 9013 (sala 34)
Telefone: 55 (0xx85) 3366 9917 (lab.)
FAX: 55 (0xx85) 3366 9450
http://www.fisica.ufc.br/raiosx
VoIP (Skype): marcossasaki
E-mail: [EMAIL PROTECTED], [EMAIL PROTECTED]



Re: advice on new powder diffractometer

2008-02-15 Thread Jon Wright

Kurt Leinenweber wrote:

... with wRp of 25% and crystallographic R also near 25%  ..


Maybe the background is too low :-?

Jon




RE: advice on new powder diffractometer

2008-02-15 Thread Von Dreele, Robert B.
Hi Kurt,
If you plot the results in POWPLOT  do the statistical analysis what does 
the delt/sig vs 2-theta plot look like? If weights are right then the band of 
delt/sig ought to be roughly 2-3 sig wide. Is your data in cpm? Some 
instruments make that by default. That will mess up the R-factors.
Bob



From: Kurt Leinenweber [mailto:[EMAIL PROTECTED]
Sent: Fri 2/15/2008 12:44 AM
To: rietveld_l@ill.fr
Subject: advice on new powder diffractometer



Hi All,

I am attempting to assist others to choose a new powder x-ray
diffractometer.  We are currently looking at a Panalytical X'Pert Pro in
alpha-1 configuration, and I would like to know if anyone has any
experience with this for Rietveld refinements?  For other reasons we
would like to keep the sample flat, but then we could not use the
alpha-1 configuration, so instead we would consider a hybrid
monochromator - I would like to ask if this monochromator makes it
impossible to do Rietveld?  Also, any other insights and suggestions for
systems that are competitive with these would be welcome.  We are at an
early stage and have time to think about this, will be looking at other
possibilities, and I would appreciate any suggestions.

Incidentally, I received some data on quartz (supplied by me) from the
alpha-1, and refined it using GSAS.  The refined structural and thermal
parameters are fine, but I get strange r-factors, with wRp of 25% and
crystallographic R also near 25%, chi**2 of 0.5.  Does anyone know why
this might happen?  With my old Siemens D5000 I easily get down to 10%
or better in the r-factors right away, so something obvious must be
wrong.

Thank you,

- Kurt Leinenweber










PS RE: advice on new powder diffractometer

2008-02-15 Thread Kurt Leinenweber
Dear Rietvelders:

P.S.  I should mention that with that background of only 2 counts, the
maximum signal was around 2000 counts.

Thanks to everyone for all the help so far.

- Kurt

-Original Message-
From: Jon Wright [mailto:[EMAIL PROTECTED] 
Sent: Friday, February 15, 2008 4:05 AM
To: rietveld_l@ill.fr
Subject: Re: advice on new powder diffractometer

Kurt Leinenweber wrote:
 ... with wRp of 25% and crystallographic R also near 25%  ..

Maybe the background is too low :-?