Hi Sun

Your bond length & angle RMSD's look suspiciously high for a 2.8 Ang
structure; this usually means that some weighting parameter(s) is/are
not optimal.  2.8 Ang is not that far from the point where the optimal
choice of structure parameters may be torsion angles instead of
Cartesian co-ordinates, in which case the optimal RMSD's for bond
lengths & angles would be exactly zero.

You should be optimising all weights that have been set arbitrarily
by the program (i.e. not obtained from independent experimental
sources), this includes not just the X-ray weight but also the B-factor
restraint weight(s) (the usual culprit) and the NCS restraint weight(s),
as Clemens suggests.  I now use the free log(likelihood) to optimise
the weights rather than Rfree, this is now printed by newer versions
of Refmac, but it's up to you which you believe (the difference in the
results may not be significant anyway).  Alternatively CNS &
phenix.refine have scripts which automatically optimise the weights
(against Rfree) for you (maybe one day CCP4/Refmac will have this very
useful capability ;-) ).

Also I would check your waters manually - don't believe everything
the auto-water placing software tells you, i.e. does the density
look sensible (at least roughly spherical shaped), is it possible
they are something other than water (check for excess density
and/or suspiciously low B factor), do they all H-bond to protein
and/or other waters you are confident about.  I had a structure at
2.9 Ang where I found only 10 good waters and that was for 900 residues
in the a.u. (maybe it had something to do with the fact that the solvent
content and average B were quite high and the data was partially twinned
so the map quality was poor). I'm sure others have opinions on how many
waters you expect to find at various resolutions.

HTH

-- Ian

> -----Original Message-----
> From: Sun Tang [mailto:[EMAIL PROTECTED] 
> Sent: 04 February 2008 22:32
> To: Ian Tickle
> Cc: CCP4BB@JISCMAIL.AC.UK
> Subject: RE: [ccp4bb] an over refined structure
> 
> Hi Ian,
> 
> Thank you very much for your detailed information.
> 
>  I checked the effect of weighter term (wa) in CCP4i for the 
> R/Rfree. When I used  wa=0.01 ,  the  value is 0.225/0.277 
> FOM =0.799.  The values  changed to 0.204/0.269  (FOM=0.806) 
> for  wa= 0.05, 0.195/0.268 (FOM=0.807) for wa=0.1 and 
> 0.186/0.267 (FOM=0.807) for wa=0.2, respectively. It seemed 
> that increase in wa decreases both R and Rfree with R more 
> than Rfree. 
> 
> Which wa value is the best one in this case?
> 
> Thank you very much for your valuable help.
> 
> Best,
> 
> Sun
> 
> Ian Tickle <[EMAIL PROTECTED]> wrote:
> 
> 
>       Hi Sun Tang
>       
>       Unfortunately there's no such thing as a fixed value 
> for the maximum acceptable Rfree-Rwork difference that 
> applies in all circumstances, because the 'normal' difference 
> depends on a number of factors, mainly the 
> observation/parameter ratio, which depends in turn on the 
> resolution and the solvent content (a greater solvent content 
> means a bigger cell volume which means more reflections for a 
> given number of ordered atoms in the a.u. and hence a bigger 
> obs/param ratio). The Rfree-Rwork difference also depends on 
> Rwork itself (i.e. you tend to get higher values of 
> Rfree-Rwork for higher values of Rwork), so it's better to 
> think in terms of the Rfree/Rwork ratio (which is independent 
> of Rwork).
>       
>       So for example at very high resolution a 'normal' value 
> for Rfree-Rwork might be only 0.02 (so 0.05 which is what 
> many people consider acceptable would actually be 
> unacceptably high), whereas at low resolution it might be 0.1 
> (so 0.05 would be unacceptably low). Also you need to bear in 
> mind that Rfree tends to have a quite high uncertainty, 
> particularly at low resolution (because it's usually based on 
> a relatively small number of observations), so the deviation 
> has to be quite big (e.g. > 3 SU) before it can be considered 
> to be statistically significant.
>       
>       So Rfree needs to be compared not with Rwork at all but 
> with the value of the optimal Rfree/Rwork expected on the 
> basis that the model parameterisation and weighting of X-ray 
> terms and restraints are optimal and the errors in the model 
> have the same effect as the random experimental errors in the 
> data (i.e. a statistical 'null hypothesis'). As Tim just 
> pointed out we tried to do this in our Acta D (1998) papers: 
> there you can compare your observed Rfree/Rwork ratio either 
> with the theoretical value or with the value found for 
> 'typical' structures in the PDB at the same resolution.
>       
>       An abnormal Rfree/Rwork ratio could arise from a number 
> of causes, not just over-fitting (I assume that's what you 
> mean by 'over-refinement' - it's not clear to me how a 
> structure can be 'over-refined' since a fundamental 
> requirement of the maximum likelihood method is that the 
> structure is always refined to convergence, and refining 
> beyond that will by definition produce no further 
> statistically significant changes in the parameters).
>       
>       For example the number of parameters being refined may 
> be either too low, or too high (over-fitting), or the values 
> of the weighting parameters may not be appropriate, or there 
> may be something badly wrong with the atomic model (e.g. 
> mistraced chain). Given the values you are reporting I think 
> the latter is very unlikely, possibly you just need to tweak 
> the X-ray and/or restraint weights.
>       
>       HTH
>       
>       Cheers
>       
>       -- Ian
>       
>       > -----Original Message-----
>       > From: [EMAIL PROTECTED] 
>       > [mailto:[EMAIL PROTECTED] On Behalf Of Sun Tang
>       > Sent: 04 February 2008 16:56
>       > To: Boaz Shaanan
>       > Cc: CCP4BB@JISCMAIL.AC.UK
>       > Subject: Re: [ccp4bb] an over refined structure
>       > 
>       > Hi Boaz, 
>       > 
>       > Thank you for your opinions. The resolution is 2.8A and I 
>       > remembered some people may think the structure is 
>       > over-refined when the difference between Rfree/Rwork is 
>       > greater than 6. 
>       > 
>       > What do you think the greatest acceptable difference 
> between the two?
>       > 
>       > Best,
>       > 
>       > Sun 
>       > 
>       > Boaz Shaanan wrote:
>       > 
>       > Hi,
>       > 
>       > Why do you think this structure is over-refined ? The 
>       > Rfree/Rwork difference of 6.2% seems fine, although you 
>       > didn't mention resolution. If anything, an over-refined 
>       > structure would show a smaller difference, as far as I know. 
>       > If all the other criteria (Ramachandran outliers, etc., map) 
>       > are OK you should just be happy with your structure.
>       > 
>       > Cheers,
>       > 
>       > Boaz
>       > 
>       > ----- Original Message -----
>       > From: Sun Tang 
>       > Date: Monday, February 4, 2008 18:41
>       > Subject: [ccp4bb] an over refined structure
>       > To: CCP4BB@JISCMAIL.AC.UK
>       > 
>       > > Hello All,
>       > > 
>       > > I refined a structure with Refmac in CCP4i and the 
> R/Rfree is 
>       > > 0.215/0.277. The difference between R and Rfree is 
>       > too much even 
>       > > though I used 0.01 for weighting term in the 
> refinement (the 
>       > > default value is 0.3). The RMSD for bond length and 
>       > bond angle 
>       > > is 0.016 A and 1.7 degree. 
>       > > 
>       > > What may be wrong with the over-refined structure? 
>       > What is the 
>       > > reason for leading to an over-refined structure? How 
>       > to avoid it?
>       > > 
>       > > Best wishes,
>       > > 
>       > > Sun Tang
>       > > 
>       > > 
>       > > ---------------------------------
>       > > Be a better friend, newshound, and know-it-all with Yahoo! 
>       > > Mobile. Try it now.
>       > 
>       > 
>       > Boaz Shaanan, Ph.D. 
>       > Dept. of Life Sciences 
>       > Ben-Gurion University of the Negev 
>       > Beer-Sheva 84105 
>       > Israel 
>       > Phone: 972-8-647-2220 ; Fax: 646-1710 
>       > Skype: boaz.shaanan
>       > 
>       > ââ'¬Å½ 
>       > 
>       > 
>       > ________________________________
>       > 
>       > Never miss a thing. Make Yahoo your homepage. 
>       > 
>       > 
>       
>       
>       Disclaimer
>       This communication is confidential and may contain 
> privileged information intended solely for the named 
> addressee(s). It may not be used or disclosed except for the 
> purpose for which it has been sent. If you are not the 
> intended recipient you must not review, use, disclose, copy, 
> distribute or take any action in reliance upon it. If you 
> have received this communication in error, please notify 
> Astex Therapeutics Ltd by emailing 
> [EMAIL PROTECTED] and destroy all copies of the 
> message and any attached documents. 
>       Astex Therapeutics Ltd monitors, controls and protects 
> all its messaging traffic in compliance with its corporate 
> email policy. The Company accepts no liability or 
> responsibility for any onward transmission or use of emails 
> and attachments having left the Astex Therapeutics domain. 
> Unless expressly stated, opinions in this message are those 
> of the individual sender and not of Astex Therapeutics Ltd. 
> The recipient should check this email and any attachments for 
> the presence of computer viruses. Astex Therapeutics Ltd 
> accepts no liability for damage caused by any virus 
> transmitted by this email. E-mail is susceptible to data 
> corruption, interception, unauthorized amendment, and 
> tampering, Astex Therapeutics Ltd only send and receive 
> e-mails on the basis that the Company is not liable for any 
> such alteration or any consequences thereof.
>       Astex Therapeutics Ltd., Registered in England at 436 
> Cambridge Science Park, Cambridge CB4 0QA under number 3751674
>       
>       
>       
> 
> 
> ________________________________
> 
> Never miss a thing. Make Yahoo your homepage. 
> <http://us.rd.yahoo.com/evt=51438/*http://www.yahoo.com/r/hs>  
> 


Disclaimer
This communication is confidential and may contain privileged information 
intended solely for the named addressee(s). It may not be used or disclosed 
except for the purpose for which it has been sent. If you are not the intended 
recipient you must not review, use, disclose, copy, distribute or take any 
action in reliance upon it. If you have received this communication in error, 
please notify Astex Therapeutics Ltd by emailing [EMAIL PROTECTED] and destroy 
all copies of the message and any attached documents. 
Astex Therapeutics Ltd monitors, controls and protects all its messaging 
traffic in compliance with its corporate email policy. The Company accepts no 
liability or responsibility for any onward transmission or use of emails and 
attachments having left the Astex Therapeutics domain.  Unless expressly 
stated, opinions in this message are those of the individual sender and not of 
Astex Therapeutics Ltd. The recipient should check this email and any 
attachments for the presence of computer viruses. Astex Therapeutics Ltd 
accepts no liability for damage caused by any virus transmitted by this email. 
E-mail is susceptible to data corruption, interception, unauthorized amendment, 
and tampering, Astex Therapeutics Ltd only send and receive e-mails on the 
basis that the Company is not liable for any such alteration or any 
consequences thereof.
Astex Therapeutics Ltd., Registered in England at 436 Cambridge Science Park, 
Cambridge CB4 0QA under number 3751674

Reply via email to