Agree! 

I think for crystallographic use the nanodrop is perfectly okay to see if
the protein is 5mg/ml or 30mg/ml. But in fact I also do not trust our
instrument if it comes to more important issues like preparing solutions for
titrations or assays. And due to the small pathlength I do not trust
absorptions of small concentrated samples at all. I always prefer a "real"
2-beam spectrophotometer (monochromators) with a quarz-cuvette and a nice
pathlength. Of course, you cannot reliably measure solutions exceeding Abs 1
or maybe 1.5 OD in a spectrophotometer with 1cm pathlength.



There's also one quite strange thing about the nanodrop - they sell the
"calibration check solution" (which is some kind of yellow chromate-solution
with known absorption), then you check your nanodrop with it and maybe find
out that it's off to some certain extend: But then you're stuck(!), because
you cannot calibrate it on your own. I guess it would be quite easy to
integrate a calibration-option into the software, but at the moment the
instrument tells you "calibration failure" and you have to call the service
guys who then carry it home and calibrate it by turning of the two small
screws at the top of it and then glue them with locktite.

Anyway, at least for our mid to high concentrated samples the nanodrop is
not showing large fluctuations so we are happy with it. But everyone using a
nanodrop should check it from time to time - as I found out that ours was
off more than 20% at one day - which raised some trouble of course.

 

Cheers,

 

Gregor

 

 

Von: CCP4 bulletin board [mailto:[EMAIL PROTECTED] Im Auftrag von Filip
Van Petegem
Gesendet: Donnerstag, 4. Dezember 2008 22:20
An: CCP4BB@JISCMAIL.AC.UK
Betreff: Re: [ccp4bb] suggestions for UV spectrometer

 

I want to add I absotely hate the nanodrop.  We've had a demo for it, and
found the readouts to be very unreliable.  Fluctuations of 20% and more.
Just leaving the same drop in and measuring the sample multiple times gives
different values (going in both directions, so not only due to
evaportations). Sure, it's easy and fast, and maybe good to have a rough
idea about your protein concentration, but I would never want to use it for
exact measurements such as needed for e.g. a CD or an ITC instrument. I've
heard other labs in our department have similar issues.  We've also had a
demo for the Nanovue from GE Healthcare:  same issues - very large
fluctuations from one sample to another.  I suppose this is simply an
inherent problem with small volumes...

Cheers

Filip Van Petegm




On Thu, Dec 4, 2008 at 12:48 PM, Patrick Loll <[EMAIL PROTECTED]> wrote:

At the risk of dragging this discussion even further afield from
crystallography:

 

How can you get realistic numbers for concentrated solutions using the
Nanodrop?  I understand that the instrument reduces absorbance by using a
very short path length. However, I thought that in order for the
Beer-Lambert formalism to be applicable, the solution needs to be
sufficiently dilute so that the chance of molecules "shadowing" one another
is negligible. Isn't this condition violated for concentrated solutions
(even with short path lengths)?

 

Pat

 

On 4 Dec 2008, at 1:27 PM, Michael Giffin wrote:





We also like the Nanodrop...

----------------------------------------------------------------------------
-----------

Patrick J. Loll, Ph. D.                                  

Professor of Biochemistry & Molecular Biology

Director, Biochemistry Graduate Program

Drexel University College of Medicine

Room 10-102 New College Building

245 N. 15th St., Mailstop 497

Philadelphia, PA  19102-1192  USA

 

(215) 762-7706

[EMAIL PROTECTED]

 




-- 
Filip Van Petegem, PhD
Assistant Professor
The University of British Columbia
Dept. of Biochemistry and Molecular Biology
2350 Health Sciences Mall - Rm 2.356
Vancouver, V6T 1Z3

phone: +1 604 827 4267
email: [EMAIL PROTECTED]
http://crg.ubc.ca/VanPetegem/