I agree with you.  If an agency disallows a particular test set-up or
procedure, they should be able to articulate that concern in writing and
justify it as well.  I know that may sound somewhat idealistic.
 
Ralph McDiarmid, AScT 
Compliance Engineering Group
Xantrex Technology Inc.
www.xantrex.com

From: Stone, Richard A (Richard) [mailto:rsto...@lucent.com] 
Sent: April 28, 2003 2:02 PM
To: 'Ralph McDiarmid'; emc-p...@majordomo.ieee.org
Subject: RE: Thermocouple glue
 
I don't agree with the general approach and opinions
on traditional yeah vs. nay....just because they ( Safety Agency) doesn't
understand,
if the Mfr'r uses it in the right context, and it does not 
promote a safety deviaition than it should be accepted...
a concrete reasoning for NOT accepting it should be pushed for
and confirmed by looking at the safety standard being used...
one man's opinion again:
Richard,

From: Ralph McDiarmid [mailto:ralph.mcdiar...@xantrex.com]
Sent: Monday, April 28, 2003 4:45 PM
To: emc-p...@majordomo.ieee.org
Subject: RE: Thermocouple glue
If it can be shown that the thermocouple remains attached and in good contact
with the surface to be measured, I don't know why the agencies would not
accept the use of thermal grease.  There concern might be the affect the
thermal compound has on the adhesive of the tape.
 
Sometimes, they oppose things they simply don't understand or have no
experience with.  Fair enough, I suppose, I can see their point of view and
they must take a conservative approach at all times.
 
Ralph McDiarmid, AScT 
Compliance Engineering Group
Xantrex Technology Inc.
www.xantrex.com

From: Brian O'Connell [mailto:boconn...@t-yuden.com] 
Sent: April 28, 2003 9:55 AM
To: emc-p...@majordomo.ieee.org
Subject: RE: Thermocouple glue
 
My use of thermal grease was discontinued several years ago by request of
various agency engineers reviewing test data/technique. And more recently,
during my ISO 17025 audit, the NCB auditor explicitly directed me to never use
thermal grease for thermocouple application. And auditors from other
NRTLs/NCBs have emphasized, at least verbally, that thermal grease is not
acceptable.
 
R/S, 
Brian 
-----Original Message----- 
From: Ron Pickard [ mailto:rpick...@hypercom.com] 
Sent: Friday, April 25, 2003 10:19 AM 
To: emc-p...@majordomo.ieee.org 
Subject: Re: Thermocouple glue 
To all, 
I'm surprised that no one hasn't mentioned this yet. 
In the past for this application, the securement that I was introduced to was
fiberglass tape and 
that white thermal grease. The tape exhibited high thermal stability and was
used to secure the 
thermocouples, but left adhesive residue when removed after a temperature
test. The thermocouple was 
inserted into the grease which offered excellent thermal conduction from the
measurement point  to 
the thermocouple. The downside to this grease, as anyone who's used this
grease would say, is that 
the grease is "messy to the extreme" and it generally could not be completely
removed from any 
surface that it came in contact with. And, it always found a way to get onto
unintended surfaces 
including clothing. But, as a plus, the thermal grease would stay put
physically over a very wide 
temperature range. 