RE: [Fwd: Re: Calibration of test equipment]

2006-03-20 Thread emc-p...@ieee.org
From: Jon Griver Sent: Monday, March 20, 2006 2:03 AM It is my understanding that it is purely the responsibility of the owner of the measuring instrument to decide on the calibration period. Hi, Jon. ISO 17025, Subclause 5.10.4.4, agrees with you. However, in dealing with various

RE: Calibration of test equipment

2006-03-20 Thread emc-p...@ieee.org
Ron - At a previous employer, we operated under CSA Category Certification Program. One of the basic requirements at the time was compliance with ISO Guide 25 (later becoming 17025). The point is that any company using the CCP that was not already compliant with ISO 17025 should have been

RE: Calibration of test equipment

2006-03-20 Thread emc-p...@ieee.org
[mailto:emc-p...@ieee.org] On Behalf Of rehel...@mmm.com Sent: 03/20/2006 1:07 AM To: emc-p...@ieee.org Subject: Re: Calibration of test equipment Mike, I have searched for this answer before. I have not found anything anywhere that requires a specific time period between calibration. Just manufacturers

RE: Calibration of test equipment

2006-03-20 Thread emc-p...@ieee.org
the extra on the calibration bill. I hope I have described this adequately well. Regards Tim From: emc-p...@ieee.org [mailto:emc-p...@ieee.org]On Behalf Of rehel...@mmm.com Sent: 20 March 2006 09:07 To: emc-p...@ieee.org Subject: Re: Calibration of test equipment *** WARNING

[Fwd: Re: Calibration of test equipment]

2006-03-20 Thread emc-p...@ieee.org
insurers would want them to take responsibility for specifying calibration periods which may be inappropriate for the actual use to which the measuring device is put. Jon Griver http://www.601help.com The Medical Device Designers' Guide to IEC 60601-1 Subject: Re: Calibration of test equipment

Re: Calibration of test equipment

2006-03-20 Thread emc-p...@ieee.org
Mike, I have searched for this answer before. I have not found anything anywhere that requires a specific time period between calibration. Just manufacturers recommendations or whatever is agreed between you and your calibration lab. The dangers of extending it beyond a year has been documented

RE: Calibration of test equipment

2006-03-17 Thread emc-p...@ieee.org
Technology Inc. From: emc-p...@ieee.org [mailto:emc-p...@ieee.org] On Behalf Of John Woodgate Sent: Friday, March 17, 2006 1:43 AM To: emc-p...@ieee.org Subject: Re: Calibration of test equipment In message e1ba0362b28ed211a1e80008c71ea30603387...@z-160-100-30-252.est.ibm.com, dated Fri, 17 Mar

RE: Calibration of test equipment

2006-03-17 Thread emc-p...@ieee.org
There have been many responses to this question regarding MRAs. However, I don't know if anyone is aware that the Canadian Standards Association (CSA) is in the process to eventually have their category certification clients at least ISO 17025 compliant. Also, as far as calibration interval, if

RE: Calibration of test equipment

2006-03-17 Thread emc-p...@ieee.org
Fax: (+44) 1245 453410 Mob: (+44) 7801 723735 From: Gordon,Ian [mailto:ian.gor...@bocedwards.com] Sent: 17 March 2006 09:12 To: 'Mike Hopkins'; emc-p...@ieee.org Cc: Tricia Rakiey Subject: RE: Calibration of test equipment Mike et al Our EMC lab is audited annually by TUV to ISO 17025

Re: Calibration of test equipment

2006-03-17 Thread emc-p...@ieee.org
In message e1ba0362b28ed211a1e80008c71ea30603387...@z-160-100-30-252.est.ibm.com, dated Fri, 17 Mar 2006, Gordon,Ian ian.gor...@bocedwards.com writes if, when it NEXT goes for calibration whether it was found to be outside the manufacturers specification or not. If it is found to be outside

RE: Calibration of test equipment

2006-03-17 Thread emc-p...@ieee.org
Mike et al Our EMC lab is audited annually by TUV to ISO 17025 and the company as a whole is audited annually by BSI and accredited to ISO 9001:2002. As we only do in-house testing the cost/benefit of full UKAS accreditation for the EMC lab isn't worthwhile. The concerns regarding calibration

RE: Calibration of test equipment

2006-03-17 Thread emc-p...@ieee.org
Hi Mike, the EN's dont usually mention much about calibration. I don't recall now any EMC standards saying that, but e.g. EN 60950-1 says how to calibrate the touch current measuring device. With regards to accreditation, the lab can assign the calibration interval as they wish but as it needs

Re: Calibration of test equipment

2006-03-16 Thread emc-p...@ieee.org
I think you're misunderstanding the function of the MRA's. MRA's allow non-EU states to designate testing and accreditation bodies on their own territory as Notified Bodies. Normally, in the absence of an MRA, a Notified Body has to be located on EU soil since EU Member States can only

Re: Calibration of test equipment

2006-03-16 Thread emc-p...@ieee.org
Mike, I'm not sure your right that CE marked products require Accredited calibrations. It is usual that ASSESSMENT bodies do. There is a difference. Assessment bodies of course want to further their own grip on the requirements...MHO... Mike Hopkins wrote on 3/16/2006, 3:31 PM: It's my

RE: Calibration of test equipment

2001-03-02 Thread Chris Maxwell
: Tuesday, February 27, 2001 2:20 PM To: 'emc-p...@ieee.org' Subject: RE: Calibration of test equipment As soon as a reference device goes out into the general lab population, it's subject to physical and electrical abuse. It may take you quite a while to notice that some device has just

RE: Calibration of test equipment

2001-02-27 Thread Brian O'Connell
As soon as a reference device goes out into the general lab population, it's subject to physical and electrical abuse. It may take you quite a while to notice that some device has just one attenuator range that's damaged (but not completely blown, just shifted a bit). As far as I'm concerned,

RE: Calibration of test equipment

2001-02-27 Thread Price, Ed
-Original Message- From: John Woodgate [mailto:j...@jmwa.demon.co.uk] Sent: Monday, February 26, 2001 1:52 PM To: Jon D. Curtis Cc: Flinders, Randall; michael.sundst...@nokia.com; c...@prodigy.net; brian.harl...@vgscientific.com; emc-p...@ieee.org Subject: Re: Calibration of test

RE: Calibration of test equipment

2001-02-27 Thread Price, Ed
...@curtis-straus.com] Sent: Monday, February 26, 2001 10:48 AM To: Flinders, Randall Cc: michael.sundst...@nokia.com; c...@prodigy.net; brian.harl...@vgscientific.com; emc-p...@ieee.org Subject: Re: Calibration of test equipment As I understand it the interpretation to have tracibility to your

Re: Calibration of test equipment

2001-02-27 Thread John Woodgate
3a9aa4c8.3df94...@curtis-straus.com, Jon D. Curtis jdc@curtis- straus.com wrote: I personally think this interpretation is overly severe, but we comply with it because we want our test reports to be accepted by authorities who think this process is reasonable. You are effectively succumbing to

Re: Calibration of test equipment

2001-02-26 Thread brent . dewitt
One method I have seen used is to have a reference instrument calibrated and then immediately use that instrument to calibrate/verify other instruments of it's type in house. After that process, the reference is placed back into standard duty until the next cal cycle. This way, at the time of

Re: Calibration of test equipment

2001-02-26 Thread Jon D. Curtis
As I understand it the interpretation to have tracibility to your national authority through equipment used only for calibration originated with NAMAS. Some other accreditors have picked it up since then. The requirement need not be that onerous. You can calibrate your own equipment traceably

Re: Calibration of test equipment

2001-02-26 Thread Flinders, Randall
Does this mean that a signal generator that is used for Radiated Immunity testing should not be used to calibrate Pre-Amps and Cables? How about Antenna Calibration? Can you use the same receiver you use on the OATS to calibrate those? I know this is a common practice with Commercial Test

Re: Calibration of test equipment

2001-02-26 Thread Lfresearch
Brian, my concern with your statement would be suggesting use of a NAMAS accredited lab Why not A2LA or other body. I know of a number of great cal labs with no accreditation Derek Walton --- This message is from the IEEE EMC Society Product

RE: Calibration of test equipment

2001-02-26 Thread Gorodetsky, Vitaly
Schleppers of the world, unite ! Vitaly -Original Message- From: Price, Ed [SMTP:ed.pr...@cubic.com] Sent: Monday, February 26, 2001 7:37 AM To: 'Brian Harlowe'; 'emc-p...@ieee.org' Subject:RE: Calibration of test equipment

RE: Calibration of test equipment

2001-02-26 Thread Michael . Sundstrom
I think there is a special requirement to keep the calibration equipment separate from the EMC equipment. In other words the calibration equipment can only be used for the calibration process and not for testing EMC. Michael Sundstrom Product Test Technician EMC Nokia Mobile Phones, Dallas

RE: Calibration of test equipment

2001-02-26 Thread Price, Ed
Brian: I calibrate my spectrum analyzer after every time I drop it. Ed -Original Message- From: Brian Harlowe [mailto:brian.harl...@vgscientific.com] Sent: Monday, February 26, 2001 12:54 AM To: 'emc-p...@ieee.org' Subject: Calibration of test equipment I seem to have hit a raw

Re: Calibration of test equipment

2001-02-26 Thread Pryor McGinnis
At the least, I would expect the test eqipment parameters/characteristics to be checked via other test equipment that is in current calibration traceable to NIST or NAMAS. This check would require a documented procedure. - Original Message - From: Brian Harlowe

Re: Calibration of test equipment

2001-02-26 Thread John Woodgate
f9d85b6af82bd4119ae800d0b769603b2c6...@kestrel.vgscientific.com, Brian Harlowe brian.harl...@vgscientific.com wrote: I seem to have hit a raw nerve by suggesting that emc test equipment should be NAMAS calibrated. But surely some cost for calibration should come into the equation. Surely no